Calibrating Tds Meter Question

  • Thread starter Snakeskins
  • Start date
  • Tagged users None
Snakeskins

Snakeskins

1,250
263
I have a Etekcity tds meter which I'll use to measure my tap water with for my soil grows but I'm confused about the testing liquid I bought. I don't know what unit of measurement to look for when adjusting the actual device.
On the bottle it says:

1500 ppM (NaCI scale)
2190 ppM (442 scale)
2930 S (microSiemens)

Does anyone know what those numbers mean and what numbers I should calibrate to?
 
Bannacis

Bannacis

1,238
163
I believe the naci scale 1500 ppm. the meter is factory calibrated to 342 ppm naci.
What does your meter read when put in solution? it should be close to the calibration liquid. they do not stray to far off cal.
 
Snakeskins

Snakeskins

1,250
263
I believe the naci scale 1500 ppm. the meter is factory calibrated to 342 ppm naci.
What does your meter read when put in solution? it should be close to the calibration liquid. they do not stray to far off cal.
Since I last calibrated, it reads like 1499 or something when using the testing liquid. The lowest the numbers will go is to the 1400s.
 