Just some of the GC limitations. We too have seen inconstant samples from iron labs.
http://cdn.intechopen.com/pdfs-wm/31531.pdf
more info on how the results are derived.
Response Factor
The size of a spectral peak is proportional to the amount of the substance that reaches the detector in the GC instrument. No detector responds equally to different compounds. Results using one detector will probably differ from results obtained using another detector. Therefore, comparing analytical results to tabulated experimental data using a different detector does not provide a reliable identification of the specimen.
A "response factor" must be calculated for each substance with a particular detector. A response factor is obtained experimentally by analyzing a known quantity of the substance into the GC instrument and measuring the area of the relevant peak. The experimental conditions (temperature, pressure, carrier gas flow rate) must be identical to those used to analyze the specimen. The response factor equals the area of the spectral peak divided by the weight or volume of the substance injected. If the technician applies the proper technique, of running a standard sample before and after running the specimen, determining a response factor is not necessary.
Worn Septum
An injection port septum should last between 100 and 200 injections. Higher injection port temperatures shorten the septum's lifespan. A leaking septum adversely affects the GC instrument's sensitivity.
If a portion of the specimen leaks back out of the septum, the amount of the specimen is not recorded. This event makes any eventual quantitative result erroneous. If air should leak into the injection port through a worn septum, the oxygen and water contained in air may skew the results. Any oxygen may react with the specimen components. If this happens, the GC instrument will provide results indicating the presence of this unintended reaction product, instead of the original compounds present in the specimen vial. Any water in the column adversely affects the GC instrument's ability to separate components.
Injection Port Temperature
The temperature of the GC injection port must be high enough to vaporize a liquid specimen instantaneously. If the temperature is too low, separation is poor and broad spectral peaks should result or no peak develops at all. If the injection temperature is too high, the specimen may decompose or change its structure. If this occurs, the GC results will indicate the presence of compounds that were not in the original specimen.
Residual Impurities
Ideally, all components of a specimen elute completely from the GC column. If any substance remains inside the column, the substance may elute during subsequent analyses with other specimens. This may result in an unexpected peak in the output. The peak produced should be broad.
Carrier Gas
If the GC instrument uses hydrogen for the carrier gas, the technician must consider whether the hydrogen may react with any of the compounds present in the specimen. If the hydrogen does react, a broad peak will result. When using a thermal conductivity detector, care should be taken as a false peak may occur if the carrier gas's thermal conductivity is in the range of the thermal conductivity of any compound in the specimen. An unstable carrier gas flow rate may produce a drifting baseline and false broad peaks. A carrier gas should be pure. Regular changing of the gas filter should prevent significant impurities.
Crucial Factors
GC analysis is highly reliable if the instrument is properly maintained, the technician follows proper procedures, and the interpretation of the results is competent. While some factors rarely affect GC analysis, some factors are absolutely essential for the use of reliable GC evidence. In all cases a technician must process a standard sample containing a verified composition identical to the presumed contents of the collected specimen. This standard sample must be processed before and after the collected specimen under identical conditions. Any output from the collected specimen that does not match the standard sample is inconclusive. If tabulated reference data exists for the relevant conditions, the specimen data must match the reference data.
If advance notice of GC testing is available, an adverse party should observe the procedure. If a retained consultant or the knowledgeable attorney observes the technician's use of the GC instrument, important information can be recorded. The technician's preparation of the specimen and the subsequent injection can be observed for errors or malfunctioning equipment. The observer should record the instrument's make, model, serial number, injection temperature, column temperature, carrier gas flow rates and pressure, identify the type of detector used, and observe any manipulation of the data by use of a computer. Ensure that the technician properly starts measuring the time at injection and records the time of elution. Any discrepancy in the time will produce an erroneous retention time. If the procedure can not be observed, the adverse party should seek all pertinent information (experimental conditions, measurements, instrument identification) and hard copy output.
http://cdn.intechopen.com/pdfs-wm/31531.pdf
more info on how the results are derived.
Response Factor
The size of a spectral peak is proportional to the amount of the substance that reaches the detector in the GC instrument. No detector responds equally to different compounds. Results using one detector will probably differ from results obtained using another detector. Therefore, comparing analytical results to tabulated experimental data using a different detector does not provide a reliable identification of the specimen.
A "response factor" must be calculated for each substance with a particular detector. A response factor is obtained experimentally by analyzing a known quantity of the substance into the GC instrument and measuring the area of the relevant peak. The experimental conditions (temperature, pressure, carrier gas flow rate) must be identical to those used to analyze the specimen. The response factor equals the area of the spectral peak divided by the weight or volume of the substance injected. If the technician applies the proper technique, of running a standard sample before and after running the specimen, determining a response factor is not necessary.
Worn Septum
An injection port septum should last between 100 and 200 injections. Higher injection port temperatures shorten the septum's lifespan. A leaking septum adversely affects the GC instrument's sensitivity.
If a portion of the specimen leaks back out of the septum, the amount of the specimen is not recorded. This event makes any eventual quantitative result erroneous. If air should leak into the injection port through a worn septum, the oxygen and water contained in air may skew the results. Any oxygen may react with the specimen components. If this happens, the GC instrument will provide results indicating the presence of this unintended reaction product, instead of the original compounds present in the specimen vial. Any water in the column adversely affects the GC instrument's ability to separate components.
Injection Port Temperature
The temperature of the GC injection port must be high enough to vaporize a liquid specimen instantaneously. If the temperature is too low, separation is poor and broad spectral peaks should result or no peak develops at all. If the injection temperature is too high, the specimen may decompose or change its structure. If this occurs, the GC results will indicate the presence of compounds that were not in the original specimen.
Residual Impurities
Ideally, all components of a specimen elute completely from the GC column. If any substance remains inside the column, the substance may elute during subsequent analyses with other specimens. This may result in an unexpected peak in the output. The peak produced should be broad.
Carrier Gas
If the GC instrument uses hydrogen for the carrier gas, the technician must consider whether the hydrogen may react with any of the compounds present in the specimen. If the hydrogen does react, a broad peak will result. When using a thermal conductivity detector, care should be taken as a false peak may occur if the carrier gas's thermal conductivity is in the range of the thermal conductivity of any compound in the specimen. An unstable carrier gas flow rate may produce a drifting baseline and false broad peaks. A carrier gas should be pure. Regular changing of the gas filter should prevent significant impurities.
Crucial Factors
GC analysis is highly reliable if the instrument is properly maintained, the technician follows proper procedures, and the interpretation of the results is competent. While some factors rarely affect GC analysis, some factors are absolutely essential for the use of reliable GC evidence. In all cases a technician must process a standard sample containing a verified composition identical to the presumed contents of the collected specimen. This standard sample must be processed before and after the collected specimen under identical conditions. Any output from the collected specimen that does not match the standard sample is inconclusive. If tabulated reference data exists for the relevant conditions, the specimen data must match the reference data.
If advance notice of GC testing is available, an adverse party should observe the procedure. If a retained consultant or the knowledgeable attorney observes the technician's use of the GC instrument, important information can be recorded. The technician's preparation of the specimen and the subsequent injection can be observed for errors or malfunctioning equipment. The observer should record the instrument's make, model, serial number, injection temperature, column temperature, carrier gas flow rates and pressure, identify the type of detector used, and observe any manipulation of the data by use of a computer. Ensure that the technician properly starts measuring the time at injection and records the time of elution. Any discrepancy in the time will produce an erroneous retention time. If the procedure can not be observed, the adverse party should seek all pertinent information (experimental conditions, measurements, instrument identification) and hard copy output.